Cannot set the brightness level to complete sensor analysis with QHY163
Posted: Wed Aug 05, 2020 4:57 am
I initially purchased a QHY163M. I could not find a brightness level that would complete all measurements of the sensor analysis evaluation. A light level that will work for one measurement is then too bright for another measurement. Here is the scenario:
Setup:
• The camera is QHY163M using an iPad app that allows me to set the brightness level from 1 to 100% in 1% increments. I have the iPad draped with a single thickness of a thin white t-shirt between it and the camera.
• I have the latest version of Sharpcap Pro (3.2.6383.0) and the newest version of the QHY drivers. I am using Windows 10.
Scenario:
Upon starting sensor analysis, the initial light level assessment wants me to keep the exposure between .1 and 1 second by changing the brightness level. I adjust the brightness level to be as close to and still under 1 second. If I dim the display too much and the exposure goes over 1 second, the ‘Proceed’ button goes dim, and I cannot continue until I increase the brightness to the point that the exposure is less than 1 second. The closest I have gotten is 991 ms.
Once I am close to 1000 ms, I press the press the ‘Proceed’ button and the e/ADU measurement begins and completes successfully. It then begins the linearity measurements. The exposure is set to .01 ms and begins to decrease until it reaches .001 ms. I then get an error message which states “the exposure time required is shorter than the minimum exposure value. Sensor analysis stopped. Restart with lower light levels.”
I can’t lower the light level because, if I do, the exposure will be over 1000 ms and will be too high when I restart the sensor analysis from the beginning. I have the same issue with light levels that produce initial exposures of 100 ms or 500 ms.
I continued trying the sensor analysis many times over several days and it suddenly started to complete successfully when the initial exposure was between 900 and 1000 ms. I have been able to complete the analysis on several attempts as long as I am within that initial exposure range. The one difference I noticed was that during the linearity measurement, the exposure increased from .01 ms (rather than decrease). So there was no risk of reaching the minimum exposure.
One item to note was that the offset was set to 51 for both the e/ADU and linearity measurements. Also. The sensor analysis has worked flawlessly with ASI290M mini and ASI224C.
I then purchased a QHY163C (color) and tried the sensor analysis. I encountered the same issues as I initially encountered with the monochrome version. I have tried it with initial exposures just over .1s, .5s and just under 1s. I have the same issue with the exposure time decreasing from .01ms until it hits the minimum exposure and I get the error.
The monochrome camera still completes the analysis as long as the initial exposure is just under 1 second.
The issue seems to revolve around the direction of the exposure change when the linearity measure starts. A decrease leads to the error while an increase will complete successfully. I did not make any changes to my process that I am aware of when the monochrome camera started to complete successfully. So I am not sure where to go from here. Any help would be greatly appreciated.
I have attached a log file for the error and a screen shot of the error
John
Setup:
• The camera is QHY163M using an iPad app that allows me to set the brightness level from 1 to 100% in 1% increments. I have the iPad draped with a single thickness of a thin white t-shirt between it and the camera.
• I have the latest version of Sharpcap Pro (3.2.6383.0) and the newest version of the QHY drivers. I am using Windows 10.
Scenario:
Upon starting sensor analysis, the initial light level assessment wants me to keep the exposure between .1 and 1 second by changing the brightness level. I adjust the brightness level to be as close to and still under 1 second. If I dim the display too much and the exposure goes over 1 second, the ‘Proceed’ button goes dim, and I cannot continue until I increase the brightness to the point that the exposure is less than 1 second. The closest I have gotten is 991 ms.
Once I am close to 1000 ms, I press the press the ‘Proceed’ button and the e/ADU measurement begins and completes successfully. It then begins the linearity measurements. The exposure is set to .01 ms and begins to decrease until it reaches .001 ms. I then get an error message which states “the exposure time required is shorter than the minimum exposure value. Sensor analysis stopped. Restart with lower light levels.”
I can’t lower the light level because, if I do, the exposure will be over 1000 ms and will be too high when I restart the sensor analysis from the beginning. I have the same issue with light levels that produce initial exposures of 100 ms or 500 ms.
I continued trying the sensor analysis many times over several days and it suddenly started to complete successfully when the initial exposure was between 900 and 1000 ms. I have been able to complete the analysis on several attempts as long as I am within that initial exposure range. The one difference I noticed was that during the linearity measurement, the exposure increased from .01 ms (rather than decrease). So there was no risk of reaching the minimum exposure.
One item to note was that the offset was set to 51 for both the e/ADU and linearity measurements. Also. The sensor analysis has worked flawlessly with ASI290M mini and ASI224C.
I then purchased a QHY163C (color) and tried the sensor analysis. I encountered the same issues as I initially encountered with the monochrome version. I have tried it with initial exposures just over .1s, .5s and just under 1s. I have the same issue with the exposure time decreasing from .01ms until it hits the minimum exposure and I get the error.
The monochrome camera still completes the analysis as long as the initial exposure is just under 1 second.
The issue seems to revolve around the direction of the exposure change when the linearity measure starts. A decrease leads to the error while an increase will complete successfully. I did not make any changes to my process that I am aware of when the monochrome camera started to complete successfully. So I am not sure where to go from here. Any help would be greatly appreciated.
I have attached a log file for the error and a screen shot of the error
John