Sensor Analysis
A very quick question: should sensor analysis be done at a room temperature or at the same temp as the light frames (e.g. -15'C). Thanks.
- admin
- Site Admin
- Posts: 13173
- Joined: Sat Feb 11, 2017 3:52 pm
- Location: Vale of the White Horse, UK
- Contact:
Re: Sensor Analysis
Hi,
it doesn't really matter - the information that the sensor analysis extracts (e/ADU, read noise, black level values) is largely unaffected by temperature.
When using the Smart Histogram to measure sky brightness, you should do that at your imaging temperature (ie -15) as the thermal noise counts as part of the background level.
Robin
it doesn't really matter - the information that the sensor analysis extracts (e/ADU, read noise, black level values) is largely unaffected by temperature.
When using the Smart Histogram to measure sky brightness, you should do that at your imaging temperature (ie -15) as the thermal noise counts as part of the background level.
Robin